pfa 100 Posted: Tue Jan 08, 2019 5:24 pm
does a larger sample greater than 800ul have any effect on the result
User avatar
craig stephenson
 
Posts: 1
Joined: Tue Jan 08, 2019 5:17 pm

Re: pfa 100 Posted: Sat Jan 26, 2019 9:20 pm
The minimum amount is 800 ul. If there is an error, the manufacturer recommends rerunning with 900 ul to rule out fibrin, and avoid air bubbles.

Our lab uses 900 ml, but if we have to share a sample or repeat, we can do with 800 ul. It's the technique that counts - you have to avoid introducing a bubble by not dispensing the blood all the way and lifting the pipette as soon as the last drop flows into the corner of the cartridge.
User avatar
Rachielle
 
Posts: 2
Joined: Wed Mar 25, 2015 4:16 am